The Orion's S-meter calibration certainly isn't anything to "write home about,"
but it is "representative" of the distant stations signal strength.
Running V1.373b5 I get the following readings using calibrated signal generator
and attenuator on 14.058 MHz:
Sig Level Preamp On Preamp Off "Standard Reading"
50 microvolts S9+11 S8.1 S9
-6dB S9 S7
S8
-12dB S8.1 S5.5
S7
-18dB S7 S4.5
S6
-24dB S5.8 S3
S5
-30dB S4.5 S2
S4
-36dB S3.1 S2
S3
-42dB S2.1 S1.8
S2
-48dB S1.8 S1.8
S1
S1.8 is about as low as the S-meter goes on this release of the firmware with
the RF gain at 100 and the AGC threshold set for .36 microvolts and the decay
set for 200dB/s. I use the "calibrated measurements" when I am trying to
understand antenna performance - particularly "front to back/side" measurements.
73, Merle - W0EWM
_______________________________________________
TenTec mailing list
TenTec@contesting.com
http://lists.contesting.com/mailman/listinfo/tentec
|
| <Prev in Thread] | Current Thread | [Next in Thread> |
|---|---|---|
| ||
| Previous by Date: | [TenTec] Typing CW, Lyle Dunlap |
|---|---|
| Next by Date: | [TenTec] STANDARDISATION OF S-METER READINGS, Merle Bone |
| Previous by Thread: | Re: [TenTec] STANDARDISATION OF S-METER READINGS, Steve Hunt |
| Next by Thread: | Re: [TenTec] STANDARDISATION OF S-METER READINGS, Ron Castro |
| Indexes: | [Date] [Thread] [Top] [All Lists] |