I can't tell from the downloadable manual, but all my life 1 ma meters have been more like 50 ohms if quality meters.

3.3 volt zeners may induce some modification of the meter reading because they tend to have a soft knee, not going into strong conduction at low current.

73, Jerry, K0CQ

On 12/15/2010 7:27 PM, pfizenmayer wrote:
Nope - the meters drop 2.38 volts at 1 ma - 2380 ohms and are fed from
fairly large series resistors so the 3.3 volt zeners are fine - in addition
the forward diode tends to damp the meters some as they are pretty undamped.

Hank K7HP

Most meter coils drop less than 0.1 volt so 3.3 volt zener's aren't much
protection. Better to use 1N4001 single junction silicon diodes, two in
parallel opposite polarity. They won't leak any current until the drop
gets up to about .55 volt.

73, Jerry, K0CQ

73 de hank K7HP