A while back, I ran across some online documentation describing the procedure that Tentec technicians used to calibrate the Centurion meters. I failed to print it out at the time, and now can't locat
Floyd, The procedure for calibrating meter indicated Ip and Ig on the 422 multi-meter is: 1. Insure that the unpowered meter is mechanically zeroed. 2. Unplug cable #4 located on the HV Rectifier-fil
Thanks for the info, Allen. I'll give that a try and see how close the meters are. Since there are adjustment pots in series with the meters for each function, the most accurate approach would probab
Allen - thought I'd bring you up to date on the metering issue here. I have it all worked out and the meters are now reading as they should. First, the 2.2 VDC was correct for the Ig metering as well
Thanks, Loyd. All interesting info. I checked with Jewell-Modutec about one year ago regarding availability of the Modutec 920051 meter, and was told that it was no longer available. BTW, Mouser show
Author: Joe Papworth via TenTec <tentec@contesting.com>
Date: Fri, 7 Oct 2016 21:26:15 -0400
Great info guys. Maybe Barry can add it to his Ten Tec wiki pages. Sent from AOL Mobile Mail Thanks, Loyd. All interesting info. I checked with Jewell-Modutec about one year ago regarding availabilit