TenTec
[Top] [All Lists]

Re: [TenTec] Centurion Trubleshooting help

To: tentec@contesting.com
Subject: Re: [TenTec] Centurion Trubleshooting help
From: Frank Sweeney <k9ic.ars@gmail.com>
Reply-to: Discussion of Ten-Tec Equipment <tentec@contesting.com>
Date: Fri, 27 Apr 2018 11:42:55 -0400
List-post: <mailto:tentec@contesting.com>
Thanks Floyd and Gary,
I should have mentioned that this is a Black face Centurion Purchased in
2011.

Problem is now resolved.  There was 2 things I repaired last night before I
turned off the lights

1.  C11 on the HV board was hanging, I supect that it never was solder from
the factory, I seriously doubt that was the issue
2.  reflowed solder on Q1 (I believe I had a cold solder connection)

for future reference, Bias voltage at K is 28V unkeyed, 8.5V Keyed


Note to self:  If grid current is flowing with no excitation, pull tubes
first and then check Q1 to see if short is removed.  (Not wanting to
believe that I was going to spend $100+ on a 3-500Z I went the cheap route
and pulled Q1.)  Had I not removed Q1 on day 1 I could have saved myself
several hours.

However, what a great learning experience.
1. I now understand how every circuit in the Centurion works.
2.  Ten-Tec did not update their schematics for instance my schematic in
the supplied manual shows a 2N3055, I have a MJ15001.  There are various
resistors on the HV supply that have been changed while others removed
3.  Connector 21 on the filament board is connector 3 on the HV board and
is marked Standby/Op switch on the schematic

Frank

On Thu, Apr 26, 2018 at 10:23 PM, Frank Sweeney <k9ic.ars@gmail.com> wrote:

> Ten Tec Centurion
>
>
>
> Initial problem Description:  occasionally on initial Turn, amp will not
> drive  No Grid or Plate current, high swr back to xcvr.  Gentle tap near
> relays fixed the problem.
>
>
>
> Secondary problem description:
>
> Within 2 weeks of daily use with initial problem the amp now exhibits
> about 10 mA grid current with no excitation, finals glowing and getting
> hot.  This condition went un-noticed for approximately 10 minutes.
>
>
>
> Trouble shooting:  1. DMM shows short of Q1 Bias in circuit, upon
> removing Q1 (PITA) found Q1 and all other active parts to be ok in the bias
> circuit   Short still present and pulled tubes and found F-G short in one
> tube
>
>
>
> Installed new 3-500
>
>
>
> 3rd problem description: the amp will not drive, No Plate current, No
> Grid current High SWR back to the xcvr
>
>
>
> Trouble shooting:  using external supply to power the QSK board, check
> switching,  Relays are switching,  Output relay when closed reduces
> received signal.  Input relay is moving a signal from my signal generator
> through the input filters to the cathodes as measured on my Oscilloscope.
>
>
>
> The tubes are acting like they are biased into cutoff. Retest Bias circuit
> components al test OK
>
>
>
> Replaced ICs in QSK board even though the QSK board appeared to be working
> correctly, added IC sockets and 15V transorbs while I had it apart.  Did
> not fix problem
>
>
>
> Bias resistors test good
>
>
>
> Help what else can it be?
>
> How do I measure the bias voltage?  Is it between ground and terminal K
> or is it between connector 22 and Terminal K?
>
>
>
> What voltage should I read at terminal K of the filament board during
>
> 1.      Stand-by
>
> 2.      Transmit
>
>
>
> Thanks in advance for any help
>
>
>
> 73,
>
> Frank K9IC
>
>
_______________________________________________
TenTec mailing list
TenTec@contesting.com
http://lists.contesting.com/mailman/listinfo/tentec

<Prev in Thread] Current Thread [Next in Thread>