Yes, if you will read the manual, you'll find that you switch to some band
and turn the preselector all the way one way.  That lets you receive WWV at
10 MHz.  Then you set the counter timebase so the readout ends in 000.0.
This is from dim memory but I'm sure it's in the book.

73,, Mike N4NT

----- Original Message ----- 
From: "Adam Nathanson" <adam at n4ekv.com>
To: <tentec at contesting.com>
Sent: Monday, August 18, 2003 7:02 PM
Subject: [TenTec] 544 freq display calibration procedure?


> My 544's freq display is 200Hz low.  Is there a simple way to adjust it?
>
>
> 73,
>
> Adam Nathanson