Some care has to be taken when interpreting the LP-100A's power readings.
On my version it indicates Pfwd when the "F+R" mode is selected, and
Pfwd-Pref when the "Net" mode is selected. I got caught out on several
occasions assuming the "F+R" mode meant Pfwd+Pref - it doesn't!
Note also that the derivation of Pfwd and Pref from the Voltage and
Current sensors assumes a 50 Ohm Zo; I don't know any way to alter the
characteristic impedance that the LP-100A assumes, unless it's buried
deep in the menus and I haven't found it?
Steve G3TXQ
On 22/07/2014 13:03, Carl Moreschi wrote:
> The LP-100A has two connections to the sensor. One is for current and
> one is for voltage. Power is simply the product of voltage times
> current. So for any load impedance, power will always be calculated
> correctly.
>
> Most less expensive meters only measure voltage. They calculate power
> as Voltage squared divided by 50. They always assume the load is a
> pure 50 ohms. But with a 1.5 swr, you could have a 75 ohm load or a
> 33 ohm load or a complex load with both real and imaginary
> components. If you have a 33 ohm load, your power will read high. If
> you have a 75 ohm load, your power will read low. The point is unless
> the swr is exactly 50 ohms, your power measurement will be wrong by a
> lot.
>
>
>
> Carl Moreschi N4PY