There's a software procedure in my Omni V.9 firmware that allows you to
exactly calibrate it.
Here's a clip form the Omni V.9 manual that tells how to do it.
> Crystal Calibration Mode -
>
> There is now a way to correct for crystal aging. To calibrate
> a specific band, proceed as follows:
>
> 1) Tune to the band and frequency of the known source signal.
> Mtune must be off for this procedure.
> 2) Tune the dial to read the exact frequency desired.
> 3) Press SHIFT TUNE.
> 4) Tune the signal in as accurately as possible. The display
> should not change while you are tuning. Take care to not
> tune the dial any more than is necessary.
> 5) When the signal is properly tuned, Press ENTER. This will
> exit the crystal calibrate procedure.
>
> Please note that the scratch pad memory and alternate VFO will be
> set to the value of the current VFO when the procedure is complete.
>
Carl Moreschi N4PY
58 Hogwood Rd
Louisburg, NC 27549
www.n4py.com
On 6/15/2016 4:34 PM, Dry Martini Holding Company😎 wrote:
>
> I have an Omni V with the N4PY V.9 upgrade. A problem has recently surfaced. On 40m for instance, when I am talking to someone on say 7115khz, I have to tune the rig to 7114.2. IOW, my freq display is reading lower than actual.
>
> On 20m I had a QSO on 14056, but my display shows 14057.
>
> What. Involved in calibrating the freq display?
>
> Tnx,
>
> Robert
> KG4KGL
>
>
> Sent from my iPhone