Why is it the pdf insists on keeping S9+60 at '0' whereas we all see
values around 240 from everybody.
What is the need for setting it at '0' ?
What is/are the effects when adjusting it to a numeric value around that
240 ??
73
Mark, PA5MW
On 1-5-2018 18:25, rfulco at suddenlink.net wrote:
> Rodney,
> There is a digital number you input for each S meter level which is done in the menu. You use the multi knob to set a nr in for each S meter reading. I found mine to be way off.
> So, I brought my orion to work and put it on our COM120 service monitor which has a high degree of calibration. You can be set for uv or DB and turn it to any level.
> .With S9 at 50mv on 14mc , I was able to "dial in" increments of 10db, and calibrate the meter .Below S9- things were little tricky, and I remember having to "spread" out the nrs. Anyway, I should have published these a long time ago. They appear to be pretty good. Yes preamp on.
>
> S0 0 S4 29 S8 53 +30 108
> S1 7 S5 36 S9 57 +40 157
> S2 14 S6 43 +10 87 +50 202
> S3 21 S7 50 +20 98 +60 253
>
> This should get you in ball park. I have a picture of the screen if needed.
>
> ..Gl Randy K5SL
>
> ---- Rodney <w3krq at dejazzd.com> wrote:
>> How do you get the s meter calb. on the Orion 1
>>
>> ---
>> This email has been checked for viruses by AVG.
>> http://www.avg.com
>> _______________________________________________
>> TenTec mailing list
>> TenTec at contesting.com
>> http://lists.contesting.com/mailman/listinfo/tentec
> _______________________________________________
> TenTec mailing list
> TenTec at contesting.com
> http://lists.contesting.com/mailman/listinfo/tentec
>
More information about the TenTec
mailing list