Tom Rauch wrote:
>>          And think about it. You get the same reflected power for the
>>          same
>> incident power if the load resistor is 100 ohms or 25 ohms.
>> Either produces a 2:1 SWR but a 100 ohm resistor would produce 1/4 the
>> current as a 25 ohm resistor.  OHM's law still applies.
>>          The meter does not precisely measure power for all load
>>          conditions
>> but it is real close  when the load resistance is close to 50 ohms.
>
>Except for that very last sentence Bill.
>
>The meter works VERY well for measuring power as long as you subtract 
>the reflected power from the forward power, as my measurement examples 
>have clearly shown. Theory says it should, and it actually does.

I'm not disputing Tom's measurement data, but the theory still is not 
convincing.

Reviewing where we've got to so far: hopefully, we all now agree that 
the capacitive coupling produces an RF voltage in the pickup loop that 
is proportional to the voltage on the main line, and the inductive 
coupling produces an RF voltage proportional to the current in the main 
line. The inductive signal reverses phase when the slug is physically 
reversed, but the capacitive signal does not.

Many other types of directional coupler such as the Bruene bridge work 
on the same principle - the only difference is in the methods used to 
get the capacitive and inductive signals, and to combine them with a 
choice of phases.

In the Bird, these two separate RF signals add or subtract at the output 
of the pickup loop, and the resultant voltage is detected by the diode. 
The instrument is designed so that when it is terminated in exactly 50 
ohms, and the slug is rotated to the 'reverse' position, the two 
voltages are exactly equal and opposite, so the meter displays zero 
reflected.  That's what gives the instrument its directional properties.

So far, so good, but...

When the slug is turned round to the 'forward' position, the two RF 
signals add in phase. We agree that one signal is proportional to 
voltage, and the other is proportional to current - but the meter 
displays the rectified SUM of these two signals. To give a resultant 
that is truly proportional to power under all circumstances, they would 
need to be multiplied - which they ain't.

That's why I still don't understand how the Bird can indicate a true 
difference between indicated forward and reflected power with all kinds 
of terminations. As I said at the start, I don't dispute Tom's 
experimental data, but I still don't think we have a solid theoretical 
backup for it.



Another point: whatever kind of theory is proposed, it has to work 
equally well for two separate cases:

Case 1.  Directional coupler terminated by a transmission line and 
mismatched load; theory based on forward and reflected waves.

Case 2. Directional coupler terminated by a lumped-component load; 
theory based on the relationship between voltage and current at a single 
point - NO transmission-line concepts to be used in this theory!

So we actually need TWO separate theories which logically MUST confirm 
each other.

Up to now we have mostly been thinking about Case 1, and getting 
seriously bogged down in transmission-line theory. IMO Case 2 is much 
easixxxx less difficult to think about, and should be a better way to 
get into the problem.

-- 