On 14 November 2013 00:48, Larry Benko <xxw0qe@comcast.net> wrote:
Dave,
I am also not aware of any UHF calibration standards.
I don't think there ever will be, as the connector has an undefined
impedance, so its hard to see how one could make any sort of standard.
I used the HP cal kit
which is an offset (both open and short) cal kit with the open also
having a
fringing capacitance.
The short does have some inductance too, though you can't enter that
on an 8753. The file here
http://na.tm.agilent.com/pna/caldefs/PNA/85033DE.htm
for the PNA does give the inductance.
The VNA removes the cal standards anomalies during
calibration giving a reference plane at the edge of the 3.5mm connector.
To be more precise, it is at the outer conductor mating plane.
I
think the offset is as you mentioned 32-33ps.
Looking at above, they are 29.243 ps on the open and 31.80799 ps on
the short, so my guess was not too far off.
I have hundreds of the RG-142
jumpers which came from some UHF aircraft diplexers and are manufactured
from the same batch of RG-142B. I did the cal with one of these cables
and
then cut the connector off one and soldered the PL-259 on. I was only
interested in SWR so losing the reference plane was not important and
could
still measure SWR.
Yes, but subject to the fact you are not just measuring the PL-259
mated pair, but the BNC->SMA adapters too.
I came into a buy of a large number of very high quality new adapters a
year
or so ago from a company who designed PCS type cell equipment and was
closing a design center. The BNC to SMA adapters were doubled up and
showed
SWR @ 500MHz below 1.05:1. I believe what I did to be accurate at
500MHz.
In that case, I think you should add that fact to your web page, as
looking at now, it is clear to me the results could be skewed by the
SMA to BNC adapters.
I've tested a fairly wide range of SMA & N adapters, mainly up to 6
GHz, but some to 20 GHz, and it is clear to me that there are
variations among them, and that it is significant, with some real
surprises - like return loss of female-female N from H+S being worst
than some really cheap adapters at 10% of the price. But the cheap
ones normally suffer from poorly cut threads, so will wear out what
they mate to if you use them a lot. I would expect variations of
PL-259 / SO-239 combinations might have far more variability, but
unlike you I have never tested PL-259s. I've measured adapters with N,
SMA, 3.5 mm, APC7 and WR90 waveguide, but never PL259 or BNC.
Larry, W0QE
Anyway Larry, it is an interesting page. I think if you add the fact
you consider the effect of the adapters insignificant, it gives it a
bit more credibility.
There is another way you could get better data, but it would need a
higher spec VNA with the TDR option
1) Measure up to a much higher frequency
2) Do an inverse Fourier Transform of the frequency domain data to get
time domain data
3) Put a window around the discontinuity in the time domain
4) Transform that back the window to the frequency domain with a
Fourier Transform
A VNA with option 010 (time domain) could do that, but to get
sufficient spatial resolution, one would need a lot higher frequency
VNA. Then one could eliminate the effect of the BNC/SMA adapter. As
the Window in the time domain would remove it.
Anyway, an interesting exercise, but I'd like to know if it repeatable
among difference combinations of connectors. That said, since the
electrical length is so small at 500 MHz, it probably is.
Dave, G8WRB
_______________________________________________