Some analyzers are better than others to produce this data. OZ7OU has
produced some excellent applications notes on the topic of VIA and VNA
calibration for S11 measurements, and has developed an excellent test
once the instrument is calibrated. This one is specifically written for
the DG8SAQ VNWA, but the concept applies universally.
http://www.hamcom.dk/VNWA/How%20to%20check%20the%20calibration%20for%20perfection%20of%20the%20VNWA%20or%20any%20other%20VNA.pdf
After calibration, a short length (1-2m) of coax is measured from very
low frequency to 100 MHz, and S11 in dB and the Smith chart are
examined. If calibration is good, S11 should always be negative and fall
smoothly with increasing frequency; the Smith trace should start at the
right edge of the outer limit of the chart,and smoothly circle around
the chart with a gradually falling radius due to increasing loss at high
frequencies.
And I completely agree with both Wes's and Paul's comments.
There is also the need to recalibrate to whatever fixture is used to
connect the line. I use a BNC female to dual male banana (Pomona)
adapter with alligator clips pushed onto the bananas, and calibrate to
the tips of the clips.
73, Jim K9YC
On 9/3/2019 1:30 PM, Paul Christensen wrote:
Wes,
"As has been suggested elsewhere in this thread, you might be tempted to try connecting the balanced line to the
unbalanced test port of the analyzer, in the belief that since it is a "floating" device it's also
"balanced". This has not worked for me."... "So the line radiates; it's part transmission line and
part antenna."
No argument there, but that doesn’t mean the CIA cannot report back a valid impedance measurement. I
have yet to find an instance where it hasn’t worked with a complex Z on a balanced line. I routinely
reverse the balanced termination on the unbalanced reflection port and achieve the same result. Moreover, I
obtain the same (or nearly the same) result when using a 1:1 Mini-Circuits RF transformer, then OSL calibrate
after the transformer. By contrast, I often cannot obtain identical results when changing line polarity when:
(1) using a "bench" type VNA device where powering comes from electrical mains; and (2) a
transformer isn’t used.
Paul, W9AC
_______________________________________________
_______________________________________________
TowerTalk mailing list
TowerTalk@contesting.com
http://lists.contesting.com/mailman/listinfo/towertalk
_______________________________________________
_______________________________________________
TowerTalk mailing list
TowerTalk@contesting.com
http://lists.contesting.com/mailman/listinfo/towertalk
|